Good Morning Eva
My Coagucheck XS result was 3.8.
Lab result is 2.5!!! Big difference!
WHAT’s you advise. My machine is about five years old
well that is rather a large difference. I would (as you are) want to investigate that.
3.8 would sort of encourage me to steer down however with the guidance of 2.5 as a "second opinion" in the data I would not.
So to clarify my understanding:
- at the Dr office you had a vein draw and you used your coaguchek
- the INR from the coaguchek was noted (either you wrote it down or just used the machine memory of that)
- when the lab results came in they notified you and you compared the two
So assuming that's what happened I would be interested in a few things to clarify this (no order this time)
- clarify exactly what the lab process is; do they use a machine like an I-Stat or a STAGO STA-R automatic coagulometer or do they do it the "from scratch" old fashioned way of centrifuging out the plasma and applying this to the traditional reagent approach.
- I'd want to repeat the test to see if another test shows a similar result
Just for your info
from the Stago company page.
Stago employs a unique, mechanical viscosity-based detection system in all its systems offering a reliable-result guarantee, since measurements are not affected by optical interference. In addition, this detection system offers optimum sensitivity for weak clot detection.
its worth noting that the Coaguchek does not use optical methods for determining coagulation. The CoaguChek uses "amperometric electrochemical" methods for determining the time to declare the point of "coagulation" to stop the timer. A good summary of this and other methods is here (
link). So this may explain the differences.
Its interesting isn't it, how many ways to measure a piece of string are there?
Of course its too early to call this but the development of Antiphospholipid syndrome could be the cause of a prolonged INR result (as you seem to be seeing). I would also ask your Dr about this possibility and test for that.
I would not alter dose at this stage because we don't know the answers to the above questions.
Some additional readings on the topic
from
this link:
There is a warning on the pack inserts for the Coaguchek XS testing strips which states:
“Antiphospholipid antibodies (APA) such as the Lupus Anticoagulant (LA) may falsely prolong coagulation times ie. they may cause false high INR values and false low values. Where APA are known to be present, it is imperative that a result be obtained using an APA-insensitive laboratory method for comparison”.
https://www.mayoclinic.org/diseases...holipid-syndrome/symptoms-causes/syc-20355831