Back to the leads, here is a very thorough AHA safety statement about MRI and all sorts of cardio devices and implants:
http://circ.ahajournals.org/content/116/24/2878.full
The type of lead matters, specifically temporary epicardial leads are deemed much safer than either temporary or permanent transvenous leads. The safety statement is very long, but here are a few quotes:
"Although the theoretical risk exists that MR examination in patients with retained temporary epicardial leads, which consist of electrically conductive material, could lead to cardiac excitation or thermal injury, such retained leads are typically relatively short in length, usually do not form large loops, and are generally not believed to pose a significant risk during MR scanning...Patients with retained temporary epicardial pacing wires are believed to be able to safely undergo MR procedures, and patients do not need to be routinely screened for the presence of such wires before scanning."
"Retained transvenous pacemaker and defibrillator leads (leads left in the body after explantation of the permanent pacemaker or ICD generator) pose significant theoretical risks, including heating and cardiac excitation. Retained fractured leads may pose a particularly high risk of thermal injury... MR examination of patients with retained transvenous leads is discouraged, and MR examination should only be considered in centers with expertise in MR and electrophysiology, and only in cases in which there is a strong clinical indication. MR examination should not be performed in patients with known retained transvenous leads that have fractures."
http://circ.ahajournals.org/content/116/24/2878.full
The type of lead matters, specifically temporary epicardial leads are deemed much safer than either temporary or permanent transvenous leads. The safety statement is very long, but here are a few quotes:
"Although the theoretical risk exists that MR examination in patients with retained temporary epicardial leads, which consist of electrically conductive material, could lead to cardiac excitation or thermal injury, such retained leads are typically relatively short in length, usually do not form large loops, and are generally not believed to pose a significant risk during MR scanning...Patients with retained temporary epicardial pacing wires are believed to be able to safely undergo MR procedures, and patients do not need to be routinely screened for the presence of such wires before scanning."
"Retained transvenous pacemaker and defibrillator leads (leads left in the body after explantation of the permanent pacemaker or ICD generator) pose significant theoretical risks, including heating and cardiac excitation. Retained fractured leads may pose a particularly high risk of thermal injury... MR examination of patients with retained transvenous leads is discouraged, and MR examination should only be considered in centers with expertise in MR and electrophysiology, and only in cases in which there is a strong clinical indication. MR examination should not be performed in patients with known retained transvenous leads that have fractures."