Dr. Thomas This should not effect value of CoaguChek. Ask the lab how linear are
they at INR values above 5 or 6 from separate blood draws from the same
individual. The CoaguChek is accurate and precise at the INR values of
importance, in or near the therapeutic range. All systems become non
linear at higher levels. Jack Ansell MD
[email protected] wrote:
>
> Dear Dr. Ansell, They have discontinued use of Coaguchek capillary INR testing at my KP clinic Falls Church VA. They state that Coaguchek gives nonlinear results at the higher levels. Does this mean the Coaguchek results are too high or too low at the higher levels? Is this "non-linearity" clinically significant? What is the INR range in which the Coaguchek is closest to a central lab veinstick ?Thanks
>
> --
> Martin A. Thomas MD,FACR
they at INR values above 5 or 6 from separate blood draws from the same
individual. The CoaguChek is accurate and precise at the INR values of
importance, in or near the therapeutic range. All systems become non
linear at higher levels. Jack Ansell MD
[email protected] wrote:
>
> Dear Dr. Ansell, They have discontinued use of Coaguchek capillary INR testing at my KP clinic Falls Church VA. They state that Coaguchek gives nonlinear results at the higher levels. Does this mean the Coaguchek results are too high or too low at the higher levels? Is this "non-linearity" clinically significant? What is the INR range in which the Coaguchek is closest to a central lab veinstick ?Thanks
>
> --