Research available at the Roche site for professionals, and I believe also on the ITC ProTime site show independent lab reports comparing lab readings to meter readings. The CoaguChek XS will report higher INRs a bit above the lab results -- for example, a 3.5 on a lab report may be reported as a 3.6 or 3.7 on the meter. Lower values (in the 2.0-3.0 range) don't show the difference. I'm not sure of the specific values where this difference shows up, and haven't researched this recently, but I think the InRatio also reports the higher values a bit higher than the lab results. The ProTime, which uses a different method for testing, doesn't have this issue.
The point here is that, while these tests can be extremely useful guidelines for dosing, when a meter shows a value that is out of range, it may look MORE out of range than it actually is - and it may cause you or whoever is managing the anticoagulation to overcompensate. (If your meter shows a value that seems way out of range, a lab test with a blood draw can be used to check the result your meter reported).
It may be good to keep this slight difference from lab results and meter results at higher INRs in mind when considering how to manage slightly out of range INR.