Hi
firstly I want to say that I agree with you, its encouraging, then I want to follow that up that the spirit of this reply is for any lurkers and those still "in the waiting room". It is not intended to be attempting to be a rebuttal of your points.
Over 5 years following valve implantation, "the RESILIA-tissue valves exhibited significantly reduced levels of structural hemodynamic valve deterioration compared with the contemporary valves." The authors acknowledged "meaningful limitations" of the study and "longer-term investigation is necessary,
this sort of duration and freedom from SVD goes back to porcine valves in the 1970's
https://pmc.ncbi.nlm.nih.gov/articles/PMC1767707/
The Edinburgh heart valve study of 533 patients, who had their valve implant(s) between 1975 and 1979, now reports comparative clinical outcome for mechanical versus bioprosthetic valves at 20 years.
1 The present report supplements a 12 year follow up, published in 1991.
2 The original study was prospective and randomised. The study design was modified in January 1977 in those patients randomised to receive a porcine bioprosthesis. Initially, the porcine valve used was the Hancock prosthesis, but, after January 1977, the Carpentier-Edwards valve was used because of its “substantial cost advantage”.
So, yes, I agree that to making meaningful determinations will take much longer than 5 years.
For example we know that in statistics there are averages and standard deviations and of course outliers too
So somebody exists on that 4 standard deviations point (at both ends). That'd be this guy
https://pmc.ncbi.nlm.nih.gov/articles/PMC10449611/
I think that its tempting to believe "huge strides" have been made in bioprosthetic valves; however the data shows "subtle but encouraging improvement" more around the "diversity" of the range than the maximum durability.
https://www.jtcvs.org/article/S0022-5223(19)38753-7/pdf
I thought I'd note some interesting points in that
Patient survival rates, with operative mortality excluded, were 80% ±1.7%
(standard error) at 5 years and 68% ± 2.7% at 10 years. Survival rates for
patients with aortic valve prostheses were 78% ± 2.8% at 5 years and 57% ± 5.4% at 10 years; for patients with mitral valve prostheses, survival rates were 80% ±2.2% at 5 years and 69% ±3.2% at 10 years. Freedom from thromboembolism for aortic valves was 93% ± 1.4% at 5 years and 88% ± 2.6% at 10 years; for mitral valves the freedom from thromboembolism was 89% ± 1.5% at 5 years and 84% ± 2.2% at 10 years. Freedom from degeneration or primary tissue failure for aortic valves was 97% ± 1.3% at 5 years and71% ± 7.6% at 10 years; for mitral valves these figures were 96% ± 1.2% at 5 years and 71% ± 4.1% at 10 years.
Valves in patients 35 years of age and below had a significantly greater rate of
degeneration (p < 0.001).
After 12 years' experience the porcine bioprosthetic valve has performed well with regard to patient survival and low rate of thromboembolism.
For patients older than 35 years the freedom from primary tissue failure is 80% at 10 years.
As a casual reader (not just the person to whom I'm replying) this remains pretty close to present observations.
For instance we have this thread:
https://www.valvereplacement.org/threads/different-strokes.889520/#post-931857
where the poster had:
Aortic valve replaced 7 August 2019, aged 61, with a 23mm Edwards Inspiris Resilia aortic valve via mini sternotomy at St Thomas’ Hospital, London.
Whats despite the projections of others upon me I am not a 'hater', I myself have written here on numerous occasions that the Resilia is a good choice within bioprosthetics.
well firstly the resilia isn't a bad choice and you could (depending on activity) get 20 years from it.
So I want to second your vote for your choice of a resilia
but equally I am well known for the view that a mechanical valve and well handled anticoagulation (INR) is the best chance you have for an event free post surgical life.
Usually those who are firmly "tribal" about this push back or outright (ongoingly) attack me for my data driven stance.
Best Wishes