They are refering to checking your machine against one of theirs for accuracy, but the truth is, you can't do it like that. Reagents are different, equipment is different, labs are different etc.
ITC Protime and Hemosense INRatio units have controls integrated into the test cuvettes and strips. You need not worry about these being accurate. The older Coagucheck required that you run controls, they supplied at your cost, with each box of new strips and that's how you adjusted for accuracy with them. I'm not sure, but I believe the new Coaguchek units also have them integrated so you no longer have to worry about it. Basically, what your being told is no longer valid with todays new units. I'm not even going to say it was valid before.
From Hemosense:
Do I need to run external controls to check for accuracy?
There are no external controls that need to be used separately. The two on board controls on each strip are testing the integrity of the strip and sample each time a test is run.
How closely should the INRatio correlate to my current lab?
The INR value depends on many factors. For more information please see the Technical Bulletins. However as per the ISO 17593 draft titled “Clinical laboratory testing and in vitro diagnostic test systems in vitro monitoring systems for anticoagulant therapy”: Below INR 2.0, 90% of results shall be within +/- 0.5 INR, and between 2.0 to 4.0, 90% of results shall be within +/- 30%. back to top