http://icvts.ctsnetjournals.org/cgi/content/full/5/suppl_2/S187
Many studies on this page, but two I wish to pick out in particular.
Study 1, in patients after one year, seems to favourably compare the Haemodynamics of The C/E magna verses the medtronic mosaic valve.
Study 2, performed in vitro (obviously with high concentration calcium solution) seems to favour the Mosaic Ultra valve over the C/E magna in terms on anticalcification and performance when calcified, adding that both are significantly better than previous ones.
A key question to ask, if anyone knows, is what changes there have been between the old medtronic mosaic and the new ultra valve. New anticalcification process? Better Haemodynamics? These things are key points as I'm assuming that study1 is based on the old mosaic.
A pity that one valve couldn't have been better in both, as it now becomes a weighing act to choose between them. However, we could really do with seeing if we can find some independant studies to back up these claims. Please search and post, people!
http://icvts.ctsnetjournals.org/cgi/content/full/5/suppl_2/S187
231 - O ONE YEAR HAEMODYNAMIC PERFORMANCE OF THE PERIMOUNT MAGNA AND THE MEDTRONIC MOSAIC BIOPROSTHESIS IN THE AORTIC POSITION: A PROSPECTIVE RANDOMISED STUDY
M.J. Dalmau1, J.M. González-Santos1, J. López-Rodríguez1, M. Bueno1, A. Arribas2, F. Nieto2
1Department of Cardiac Surgery, Salamanca University Hospital, Salamanca, Spain; 2Department of Cardiology, Salamanca University Hospital, Salamanca, Spain
Objectives: To compare the haemodynamic performance of the Edwards Perimount Magna (EPM) pericardial xenograft and the Medtronic Mosaic (MM) porcine bioprosthesis. Comparison was performed according to the patient aortic annulus diameter.
Methods: Eighty-six patients undergoing aortic valve replacement were prospectively assigned to receive either an EPM valve (n=43) or a MM bioprosthesis (n=43). Intraoperative randomisation was performed after the aortic annulus diameter was measured with three different sizers (EPM, MM and a Hegar dilator). Patients were grouped according to their aortic annulus diameter in <22 mm (n=13), 22?23 mm (n=31) and >23 mm (n=42). Echocardiographic assessment was performed 1 year postoperatively.
Results: The mean aortic annular diameter (EPM 23.8±2.1 mm vs. MM 23.6±2.3 mm) and mean valve size implanted (EPM 22.6±2.1 mm vs. MM 23.3±2.1 mm) were comparable in both groups. The EPM-group showed significantly lower mean gradient (EPM 10.0±3.1 mmHg vs. MM 17.2±8.3 mmHg) and larger effective orifice area (EOA) (EPM 1.97±0.4 cm2 vs. MM 1.67±0.4 cm2, P<0.0001). The EPM-valve was superior with respect to mean pressure gradient and EOA in all aortic annular diameters. This difference was statistically significant in patients with an annulus diameter of 22?23 mm (EPM 9.8±2.5 mmHg vs. MM 18.2±8.8 mmHg; EPM 1.76±0.3 cm2 vs. MM 1.52±0.2 cm2) and >23 mm (EPM 9.4±3.0 mmHg vs. MM 14.1±5.6 mmHg; EPM 2.17±0.3 cm2 vs. MM 1.90±0.4 cm2). Patient-prosthesis mismacht (EAOI < 0.85 cm2/m2) was present in 26.3% (MM) vs. 8.1% (EPM) of the patients (P=0.03).
Conclusions: When the same aortic annulus diameter is taken as a reference, the EPM-valve was haemodynamically superior to the MM-bioprosthesis. The use of the EPM-prosthesis significantly reduced the incidence of PPM.
236 - O CHANGES IN HAEMODYNAMIC PERFORMANCE AND LEAFLET KINEMATICS WITH PROGRESSIVE CALCIFICATION OF PORCINE VERSUS PERICARDIAL AORTIC VALVES
P. Kleine1, O. Dzemali1, F. Bakhtiary1, C. Schmitz2, U. Steinseiffer2, B. Glasmacher2, A. Moritz1
1Department of Thoracic and Cardiovascular Surgery University Hospital, Frankfurt am Main, Germany; 2Helmholtz Institute for Applied Medical Engineering, Aachen, Germany
Objectives: In vitro testing of biological valves has only been performed using fresh valves. The following study investigates changes in haemodynamic performance and leaflet kinematics in progressively calcified pericardial and porcine aortic valve prostheses.
Methods: Edwards Perimount Magna (PM, n=5) and Medtronic Mosaic Ultra (MU, n=5) heart valves (23 mm) were investigated in a pulse duplicator (70 beats/min, Cardiac Output 5 l/min). Leaflet kinematics were visualised with a high-speed camera (3000 frames/s). Following valves were exposed to a Calcium phosphate solution at a constant pulse rate of 300/min for a total of 6 weeks. Repeated testing was performed after 1, 2, 3, 4 and 6 weeks of calcification.
Results: Initially PM demonstrated lower pressure gradients compared to the MU (9.7±1.2 vs. 14.0±1.1 mmHg), but a higher closing volume (7.2±1.4 vs. 1.8±0.4% of stroke volume) leading to an equivalent total energy loss (8.9±1.4 vs. 9.0±1.5%). PM then calcified significantly faster and more severe leading to an increase in gradients (12.6±1.4 after 6 weeks). Leaflet kinematics showed progressively longer opening and closing times for the pericardial valves (closing time PM 135±11 msec vs. MU 85±9 ms after 6 weeks).
Conclusions: Despite the new ThermaFix tissue treatment Perimount Magna pericardial valves calcified in vitro faster and more severe than Mosaic Ultra porcine valves, which demonstrated a constant performance throughout the calcification process. Leaflet kinematics showed a progressive prolongation of opening and closing times for pericardial valves leading to higher closing volume. Both valve performed superior to previously tested biological valve substitutes.
Many studies on this page, but two I wish to pick out in particular.
Study 1, in patients after one year, seems to favourably compare the Haemodynamics of The C/E magna verses the medtronic mosaic valve.
Study 2, performed in vitro (obviously with high concentration calcium solution) seems to favour the Mosaic Ultra valve over the C/E magna in terms on anticalcification and performance when calcified, adding that both are significantly better than previous ones.
A key question to ask, if anyone knows, is what changes there have been between the old medtronic mosaic and the new ultra valve. New anticalcification process? Better Haemodynamics? These things are key points as I'm assuming that study1 is based on the old mosaic.
A pity that one valve couldn't have been better in both, as it now becomes a weighing act to choose between them. However, we could really do with seeing if we can find some independant studies to back up these claims. Please search and post, people!
http://icvts.ctsnetjournals.org/cgi/content/full/5/suppl_2/S187
231 - O ONE YEAR HAEMODYNAMIC PERFORMANCE OF THE PERIMOUNT MAGNA AND THE MEDTRONIC MOSAIC BIOPROSTHESIS IN THE AORTIC POSITION: A PROSPECTIVE RANDOMISED STUDY
M.J. Dalmau1, J.M. González-Santos1, J. López-Rodríguez1, M. Bueno1, A. Arribas2, F. Nieto2
1Department of Cardiac Surgery, Salamanca University Hospital, Salamanca, Spain; 2Department of Cardiology, Salamanca University Hospital, Salamanca, Spain
Objectives: To compare the haemodynamic performance of the Edwards Perimount Magna (EPM) pericardial xenograft and the Medtronic Mosaic (MM) porcine bioprosthesis. Comparison was performed according to the patient aortic annulus diameter.
Methods: Eighty-six patients undergoing aortic valve replacement were prospectively assigned to receive either an EPM valve (n=43) or a MM bioprosthesis (n=43). Intraoperative randomisation was performed after the aortic annulus diameter was measured with three different sizers (EPM, MM and a Hegar dilator). Patients were grouped according to their aortic annulus diameter in <22 mm (n=13), 22?23 mm (n=31) and >23 mm (n=42). Echocardiographic assessment was performed 1 year postoperatively.
Results: The mean aortic annular diameter (EPM 23.8±2.1 mm vs. MM 23.6±2.3 mm) and mean valve size implanted (EPM 22.6±2.1 mm vs. MM 23.3±2.1 mm) were comparable in both groups. The EPM-group showed significantly lower mean gradient (EPM 10.0±3.1 mmHg vs. MM 17.2±8.3 mmHg) and larger effective orifice area (EOA) (EPM 1.97±0.4 cm2 vs. MM 1.67±0.4 cm2, P<0.0001). The EPM-valve was superior with respect to mean pressure gradient and EOA in all aortic annular diameters. This difference was statistically significant in patients with an annulus diameter of 22?23 mm (EPM 9.8±2.5 mmHg vs. MM 18.2±8.8 mmHg; EPM 1.76±0.3 cm2 vs. MM 1.52±0.2 cm2) and >23 mm (EPM 9.4±3.0 mmHg vs. MM 14.1±5.6 mmHg; EPM 2.17±0.3 cm2 vs. MM 1.90±0.4 cm2). Patient-prosthesis mismacht (EAOI < 0.85 cm2/m2) was present in 26.3% (MM) vs. 8.1% (EPM) of the patients (P=0.03).
Conclusions: When the same aortic annulus diameter is taken as a reference, the EPM-valve was haemodynamically superior to the MM-bioprosthesis. The use of the EPM-prosthesis significantly reduced the incidence of PPM.
236 - O CHANGES IN HAEMODYNAMIC PERFORMANCE AND LEAFLET KINEMATICS WITH PROGRESSIVE CALCIFICATION OF PORCINE VERSUS PERICARDIAL AORTIC VALVES
P. Kleine1, O. Dzemali1, F. Bakhtiary1, C. Schmitz2, U. Steinseiffer2, B. Glasmacher2, A. Moritz1
1Department of Thoracic and Cardiovascular Surgery University Hospital, Frankfurt am Main, Germany; 2Helmholtz Institute for Applied Medical Engineering, Aachen, Germany
Objectives: In vitro testing of biological valves has only been performed using fresh valves. The following study investigates changes in haemodynamic performance and leaflet kinematics in progressively calcified pericardial and porcine aortic valve prostheses.
Methods: Edwards Perimount Magna (PM, n=5) and Medtronic Mosaic Ultra (MU, n=5) heart valves (23 mm) were investigated in a pulse duplicator (70 beats/min, Cardiac Output 5 l/min). Leaflet kinematics were visualised with a high-speed camera (3000 frames/s). Following valves were exposed to a Calcium phosphate solution at a constant pulse rate of 300/min for a total of 6 weeks. Repeated testing was performed after 1, 2, 3, 4 and 6 weeks of calcification.
Results: Initially PM demonstrated lower pressure gradients compared to the MU (9.7±1.2 vs. 14.0±1.1 mmHg), but a higher closing volume (7.2±1.4 vs. 1.8±0.4% of stroke volume) leading to an equivalent total energy loss (8.9±1.4 vs. 9.0±1.5%). PM then calcified significantly faster and more severe leading to an increase in gradients (12.6±1.4 after 6 weeks). Leaflet kinematics showed progressively longer opening and closing times for the pericardial valves (closing time PM 135±11 msec vs. MU 85±9 ms after 6 weeks).
Conclusions: Despite the new ThermaFix tissue treatment Perimount Magna pericardial valves calcified in vitro faster and more severe than Mosaic Ultra porcine valves, which demonstrated a constant performance throughout the calcification process. Leaflet kinematics showed a progressive prolongation of opening and closing times for pericardial valves leading to higher closing volume. Both valve performed superior to previously tested biological valve substitutes.